[Milsurplus] The BC-221N Freq Meter calibration.
David Olean
k1whs at metrocast.net
Fri Feb 9 16:49:52 EST 2024
Hello Hue,
I am keeping my BC-221 exactly as it was made in 1942. I have been
leaving it turned on and set at 4.000 MHz for about a week. I check it a
few times every day and am amazed at how it is rock stable now. It
varies +/- about 25 or 45 Hz over many days. When I first got the
BC-221 in 2016, I noted that it had some significant long term drift and
did not behave as it does now. I did pretty much nothing to the
circuitry except replacing the bathtub caps. There is a bypass function
with Capacitors 10-1 and 10-2 in the BC-221N VFO that could cause some
long term drift. They bypass the screen and plate of the VFO tube.
Leakage in those caps would vary the voltages to the VFO and cause
significant drift. (Other BC-221s have similar caps with different IDs)
One of the failure modes of these leaky caps is that the leakage
increases with increasing internal temperatures within the cap. Then
there is the fact that an unused BC-221 can absorb moisture when stored
in damp and cold areas. My BC-221 had an unknown background, but the
constant frequency drift was telling me that there was something
happening within the unit as it was drying out.
I have a 2nd BC-221 made by Bendix in WW2, that I will try out with
some accuracy checks. It has been in a warm environment for a few years,
but has not had any bypass caps changed. It might be a good test to
evaluate it for some before and after measurements. I really believe
that long term excessive drift of the VFO components (mostly the
inductor) causes the crystal markers to go to a different portion of the
tuning curve, and that might reduce the accuracy of the calibration
book. I am still looking into that.
We take frequency accuracy for granted these days. Today, there are
even frequency police who alert you to when you are 20 or 30 Hz off of a
net frequency. I ran a signal shop in the US Army back in 1969 and
1970, and we did not have a frequency counter anywhere in our TOE gear.
We relied on signal generators and crystal calibrators entirely. Now we
only did 3rd echelon maintenance, but fast forward to today and even
Johnnie Novice gets upset when you are a few Hz removed from where you
should be. heh heh. It is expected that my 10 GHz weak signal gear
should be within about 100 Hz for best results now. Just 20 years ago,
we were tuning +/- 25 kHz and then +/- 50 or 100 kHz if we could not
find the other guy! My hat is off to those people in WW2 who relied on
the LM or BC-221 to keep those aircraft and tank radios on frequency. It
was not easy.
My BC-221 is going back on the shelf, but I am putting a note inside
outlining what has been done to it. Maybe I could drive my Central
Electronics 20A as a VFO?
Dave K1WHS
On 2/9/2024 3:22 AM, Hubert Miller wrote:
> It is a pretty simple matter to change the valves to FETs. I think 73 Magazine did the writeup, which i have somewhere.
> Change vacuum tube test equipment to FETs, yes. For any communications device, no, or just with plug-in replacements,
> is what i say.
> -Hue Miller
> ______________________________________________________________
> Milsurplus mailing list
> Home: http://mailman.qth.net/mailman/listinfo/milsurplus
> Help: http://mailman.qth.net/mmfaq.htm
> Post: mailto:Milsurplus at mailman.qth.net
>
> This list hosted by: http://www.qsl.net
> Please help support this email list: http://www.qsl.net/donate.html
More information about the Milsurplus
mailing list