If the meter needle is bent to the left, that could certainly account for
the line voltage setting problem. Unfortunately, I think that getting into
the -B meter is difficult (if I'm not confusing it with another set).
The 47 ohm parasitic suppression resistors (3 of them) are in the TV-7(*).
But I don't know of any MWO adding them to the I-177B.
In a message dated 3/25/2011 4:06:49 PM Central Daylight Time,
ianmwilson73 at gmail.com writes:
> Paul, Robert,
>
> I think this may be along the correct lines. I measured the voltage across
> the 25k series
> resistor and did some calculations. The meter position at 0.7mA - which
> should be half
> scale - isn't correct.
>
> I also note that the meter isn't on zero when the power is off, and the
> meter pointer is
> bent somewhat to the left.
>
> I will continue to work the other problem. I don't see any 47 ohm stopper
> resistors - perhaps
> these were a field modification?
>
> Thanks for the help so far!
>
> 73, ian K3IMW