I would check/clean/replace the zero adjust pot R1. If OK, I would check R33
(150K), R34 (220K) and R35 (150K). If OK, check R14 (10K), R15 (100) and
R16 (27K). Statistically, it should be fixed by now. If not, measure the voltage
across C1 (16uF/150V) - should be 65V-80V and clean the function switch
(unlikely, but what the heck).
-- Ron
________________________________
>________________________________
> From: Garey Barrell <k4oah at mindspring.com>
>To: Heathkit Reflector <heathkit at mailman.qth.net>
>Sent: Thursday, May 31, 2012 12:03 PM
>Subject: [Heathkit] Heathkit IM-18
>
>OK, I give up.........
>
>I have an IM-18 that has been working for years. Turned it on, and the meter pegs 'downscale' on
>'AC'. On 'DC-' it pegs downscale, on 'DC+' it pegs upscale. ZERO control has no visible effect.
>No combination of the CAL controls makes any visible effect. Interestingly, results are identical
>with the 12AU7 removed. DC voltages on the power supply capacitor are nominal. Whoever built it
>did a VERY nice job, beautiful solder on board and switch. Close visual examination shows not
>obvious shorts, bridges, etc.
>
>Open to ideas or suggestions..
>
>73, Garey - K4OAH
>Glen Allen, VA
>
>Drake 2-B, 2-C/2-NT, 4-A, 4-B, C-Line
>and TR-4/C Service Supplement CDs
><www.k4oah.com>
>
>
>
>______________________________________________________________
>Heathkit mailing list
>Home: http://mailman.qth.net/mailman/listinfo/heathkit
>Help: http://mailman.qth.net/mmfaq.htm
>Post: mailto:Heathkit at mailman.qth.net
>
>List Administrator: Duane Fischer, W8DBF
>** For Assistance: dfischer at usol.com **
>
>
>This list hosted by: http://www.qsl.net
>Please help support this email list: http://www.qsl.net/donate.html
>
>
>
More information about the Heathkit
mailing list