[Heathkit] TT-1 TUBE TESTER
[email protected]
[email protected]
Fri, 5 Sep 2003 10:28:26 -0500
Question about the filament/heater voltages.
I have a TT-1-A that seems to have a grave discepancy as to actual at
socket fil voltage.
for instance, when setting the line adjust to line set and fil sw to 6
v it will show about
6 v or less at tube socket, no tube inserted. plug in tube and voltage
drops to 4.3 and
re-adjusting the line set control may bring it up
to about 4.7v.
The odd thing about this is that a 6L6 will test as good with
this heater voltage,
I measured this with a socket adapter with the terminals at the side.
I do not have the complete manual which may indicate the point to measure
the 100 volts
neccesary at the primary for proper operation.
what gives???
Jack [email protected]