Good morning to all!
We have received numerous inquiries regarding SR-2000 S-Meter insensitivity.
First, the procedure for adjusting the S-Meter included in the operator's
manual is a bit lacking. The correct procedure is to turn the AGC threshold
pot, R-46 fully clockwise, zero the meter with the meter zero pot, and then
turn the S-Meter pot CCW until the meter just moves. This is a three step
process and you are finished.
Before doing anything, check the resistors R-118, 119, and 121. They are
critical, especially R-118. You can find R119 the 82K resistor directly
below L15, R121 3.3K under the plate that holds the Meter zero pot, and R118
5.6K at V8B. Disconnect one S-Meter lead and measure all three after
unscrewing the pot plate and isolating it from ground, you won't have to
unsolder anything.
Bear in mind that the AGC/S-Meter tube is a 7059 which is a 13.5V tube
running at 12.6V if all is well and 90V on the plate. After you perform the
adjustment above, remove the S-Meter tube V8A and see what the S-meter
indicates. It should be above +60Db, pegged. Whatever it reads with the
tube removed is the highest reading possible. You will probably wonder why
the meter is pegged when you first turn it on. When first fired up, the B+
will be approximately 320V until the tubes warm up and draw plate current.
Then it will drop back to 280V (The meter is driven with a bridge). Tube V8
is the one tube in the SR-2000 that must be hand selected for S-Meter
performance. If the meter does not peg with V8 removed replace it and do
the above adjustment then pull the tube and repeat until its right.
The S-Meter will be quite accurate when you are finished.
Kindest regards Jim K9AXN