The easiest way to check the crystal frequency is to use an external
general coverage receiver with an accurate readout that covers the range
of the two crystals. If you have such a receiver, set the receiver to
exactly 5 or 10 megahertz to receive WWV, choose SSB reception, and if
the BFO pitch is adjustable set it for exact zero beat. This step
calibrates your external receiver accurately. Now without changing the
BFO pitch setting tune the receiver to pick up either the 1600 or 1700
Khz xtal and the carefully tune the external receiver for zero beat and
the dial readout is your crystal frequency. Although this is not as
accurate as using a well calibrated frequency counter it is much easier
and should get you within a few 100 hertz of the actual frequency which
is plenty close for what you are doing.
If your counter is sufficiently sensitive you can pick off the signal
from the 12AT7 oscillator but many (most?) counters are not going to be
sensitive enough to pick up enough signal for counting without
excessively loading the circuit you are trying to measure. If you have
a scope that supplies a vertical signal output (most Tektronix and HP
scopes will provide this) you can use the vertical output to drive your
counter and use the scope to sample the signal off of the 12AT7 oscillator.
By the way, all is not lost if one or the other oscillators is a little
off as you can compensate during calibration. For example, it the 1600
KC oscillator is on frequency but the 1700 is actually at 1702 Khz. then
the difference between the two is 102 Khz instead of 100 and half of
this would be 51 instead of 50 so during alignment your BFO adjustment
would be calibrated for 51 Khz. at zero beat and the low IF would now be
aligned for 51.75 instead of 50.75. This will retain the same passband
characteristic and allow carriers to remain at zero beat when switching
between upper and lower sidebands. If one xtal is too far out then this
will not work because you will exceed the adjustment range of the low IF
transformers but for a small deviation it is far easier and cheaper to
do this alignment than trying to acquire a new xtal.
73, Rodger WQ9E
Jack Schrader wrote:
> Sent: Tuesday, November 20, 2007 8:51 PM
> Subject: sx101a crystals
>
>
> Another stupid question from w1jhs.
>
> I am trying to restore my sx101a per the ER articles and one of the cautions is to make sure the two crystals are on freq, (1600 & 1700).
>
> My question is, how do I do that ? I have a freq counter.
>
> 73 Jack
> ______________________________________________________________
> Hallicrafters mailing list
> Home: http://mailman.qth.net/mailman/listinfo/hallicrafters
> Help: http://mailman.qth.net/mmfaq.html
> Post: mailto:Hallicrafters at mailman.qth.net
> ----
> List Administrator: Duane Fischer, W8DBF **for assistance**
> dfischer at usol.com
> ----
> HHI Halligan's Hallicrafters International and www.w9wze.net
>
>
>
More information about the Hallicrafters
mailing list