[GreenKeys] User manual for ME-9G/U ????

scottjohnson1 at cox.net scottjohnson1 at cox.net
Fri Sep 27 20:14:44 EDT 2024


It’s pretty straight forward, on volts you have a 1K /volt side and a 20k ohm /volt, positive lead goes in whatever range you want, negative lead goes to jack above and left of the meter. Everything else is marked.  Make sure the knob that you replace the Rx scale switch is indexed properly.  It needs a D cell and three 4.5-volt batteries in series for ohms, the three 4.5V batteries can be replaced by 9 AAs in series.  To use the 5 kV multiplier at the top, connect a jumper to the 2.5V jack, the high voltage input is a banana, but make sure you use HV lead wire of you are using it.  Most of the condensed instructions are inside the lid, which also has the leads and accessories, I take it that it was missing?  Full nomenclature is TS-352.  I believe it is the predecessor to the PSM-6 that all GIs know and love.

 

Scott W7SVJ

 

From: greenkeys-bounces at mailman.qth.net <greenkeys-bounces at mailman.qth.net> On Behalf Of steve bennett via GreenKeys
Sent: Friday, September 27, 2024 12:36
To: GREENKEYS BULLETIN BOARD <greenkeys at mailman.qth.net>
Subject: [GreenKeys] User manual for ME-9G/U ????

 

Can anyone tell me how to use this meter?

Just picked it up today.

 

Thanks,

-Steve