[GPS_Standard] Fw: Fixed value offset for frequency
Stanley Reynolds
stanley_reynolds at yahoo.com
Fri Feb 13 21:05:55 EST 2009
----- Forwarded Message ----
From: Stanley Reynolds <stanley_reynolds at yahoo.com>
To: "Bert, VE2ZAZ" <ve2zaz at sympatico.ca>
Sent: Friday, February 13, 2009 8:04:55 PM
Subject: Re: [GPS_Standard] Fixed value offset for frequency
Bert,
Checking the counter ( a Racal-Dana 1992 ) as you suggest it reads a stable 5,000,000,00 with either the GPSDO or Caesium standard. Also the Caesium reads very close to .30 usec or 5X10E-12 as measured with a Austron 2100f loran receiver. The GPSDO offset is much more at 100 to 200 usec measured with the loran. The difference between the two as measured with the counter is very stable over a 24 hour period just the GPSDO is a little low. I think I will look closer at the GPSDO power supplies and RF signals. The output looks clean on the scope with a 50 ohm terminator, but I have not looked at the pps and oscillator outputs before they get to the controller. What ever is in error is very stable and doesn't drift much as I can tell with my limited skill. I also have other GPSDOs but it is all unproven.
As I was writing this I had the idea to place a tee and terminator at the counter external standard input it now reads as expected, problem solved :-)
Stanley N4IQT
________________________________
From: "Bert, VE2ZAZ" <ve2zaz at sympatico.ca>
To: gps_standard at mailman.qth.net
Sent: Friday, February 13, 2009 3:38:55 PM
Subject: Re: [GPS_Standard] Fixed value offset for frequency
Hi Stanley,
Unfortunately it is not possible to add such a small offset to the FLL.
Changing the nominal value to 0x67FF will cause a one-Hertz offset, which is
way too much.
Instead of trying to patch up your counter, I suggest you try finding what
is wrong.
A good way to verify if your counter has a problem is to feed any 5MHz
signal, say your frequency standard, to both the RF input and the reference
input. The counter should normally alternate the dispaly between 4,999,999
and 5,000,000, or always 5,000,000 depending on the type of counter. If you
still read high, then there is a problem. Also check for power supply noise
on all rails of your counter.
Assuming the counter is fine, what guarantee do you have that your Caesium
standard is within specs? Are you convinced that the GPSDO has achieved the
expected accuracy? What makes you think you can trust it? We are dealing
with "act of faith" devices here. We should always question something we
cannot demonstrate.
Let us know the outcome!
Bert, VE2ZAZ
______________________________________________________________
GPS_Standard mailing list
Home: http://mailman.qth.net/mailman/listinfo/gps_standard
Help: http://mailman.qth.net/mmfaq.htm
Post: mailto:GPS_Standard at mailman.qth.net
This list hosted by: http://www.qsl.net
Please help support this email list: http://www.qsl.net/donate.html
-------------- next part --------------
An HTML attachment was scrubbed...
URL: http://mailman.qth.net/pipermail/gps_standard/attachments/20090213/aca786fc/attachment.html
More information about the GPS_Standard
mailing list