[Elecraft] A question on the P3 or actually the Tx Mon (K3)
Alan Bloom
n1al at sonic.net
Mon Mar 6 12:34:55 EST 2017
Hi Paul,
It's just a matter of adjusting the SensorCal menu setting until the
power reads the same as the device used for the calibration. Of course,
that assumes the calibration device is more accurate than the P3. :=)
The nominal calibration value is 500. The intent was to make it
correspond to the same value as in a W2, but the match is not perfect.
For one thing, the P3 uses a more-sophisticated algorithm to match the
diode detector nonlinearity than the W2.
Alan N1AL
On 03/06/2017 08:56 AM, paul ecker via Elecraft wrote:
> Alan - is there any documentation available on the procedure for calibrating the Tx Monitor?
> 73Paul