Harlan,
That is correct - improper test leads are often a 'culprit'.
For instance trying to measure an HF signal with a 'scope. The
'scope may b rated for 200 MHz, but if you try to us a "hunk of coax"
for a probe, the frequency rating may diminish to 2 MHz (OK, that is
just an example). You need to consider the whole measurement system. A
scope's frequency rating must also be coupled with the frequency rating
of the probe. The use of 10X probes is necessary to keep capacitance
loading of the circuit under test to a minimum. And those probes have a
frequency rating too. Take the entire setup of the test equipment into
consideration when doing measurements.
Make certain that every part of the test equipment being used is
suitable for the frequency involved. A simple coax probe may be OK for
AF frequencies, but it is not adequate when the frequency is increased.
A 'scope can show proper RF voltage amplitude, but the frequency rating
of the probe and the 'scope should be taken into consideration. For
instance, a 35 MHz scope will only show a signal that is 3 dB down at 35
MHz. I would suggest that valid RF voltage measurements be made at
1/10 of the 'scope's frequency rating (and that goes for the probe too).
Again I state, trust no measurement equipment until its accuracy has
been verified.
73,
Don W3FPR
On 11/21/2015 10:09 PM, hsherriff wrote:
> I'm the supervisor of a metrology lab and that is one of the first
> things I ensure a new technician understands. And many times it isn't
> the test equipment that is at fault, nor the unit under test, but the
> methodology of using that test equipment. Not using the equipment
> "guard" correctly, induced voltages, improper test leads for the
> job.... and many many more.
>
>