[Elecraft] K3 Wattmeter Cal and Tech Mode
Bruce Beford
bruce.beford at myfairpoint.net
Mon Apr 7 17:11:24 EDT 2014
Ed,
This seems to indicate the KXV3 menu setting is set to TEST. See the
calibration procedure on pg 49 of the K3 manual, under 1.0 milliwatt meter
calibration. Set the COBFIG:KXV3 menu entry to "NOR" (or "Not Inst", if you
don't have the KXV3 or 3A), and you should be able to proceed with the LP
and HP calibrations.
GL,
Bruce N1RX
> Seems no PWR setting from 0 up to 100 watts makes any difference. I
> only get the "WMTR MW" screen in the config menu. I can't get to the LP
> or HP screens to make any adjustments. Tech Mode is turned on.