On 10/13/2013 6:22 PM, Jack Berry wrote:
> When operating I see a big difference between the noise floor and
> signal indications between the K3 meter and P3 display. For instance,
> on 80 meters tonight I see a noise floor of around S2-3 on the P3 and
> a signal peaking just over S9.
>
> At the same time the K3 meter is displaying a noise level of S9 and a
> signal level of 20 over S9.
>
> Is that a calibration error, a result of settings that I have
> selected - or just the way things are?
Well Jack, it could be a calibration error, but
1. Do you have the "K3 IF output" mod installed. It's one resistor,
and it raises the 8MHz IF output to the P3.
2. I calibrated my S-Meter with my service monitor, S9 for 50uv in.
Then I adjusted the P3 gain to give me -73dBm on the screen.
3. The P3 averages signals [you can set the sample size, 2 is the
minimum], and my P3 never shows the same reading as my S-Meter on CW or
SSB except when I'm doing RTTY or some other constant amplitude mode,
and sometimes even those aren't exactly on what I expect them to be.
4. I don't think it's a problem for you, it's just inherent in the
technology.