[Elecraft] K3: RF Calibration / RF output
Bill W4ZV
btippett at alum.mit.edu
Tue Mar 13 07:39:11 EDT 2012
Edward R. Cole wrote
>
> I obtained a brand-new calibrated 25H element for my Bird43 power
> meter, today. I checked power out on my K3/10 set to 12.0w.
> 80m = 11w
> 40m = 11w
> 20m = 11.5w
> 15m = 11.5w
> 10m = 11.5w
> all measurements into a Bird 25w coaxial termination
>
Bird accuracy is +/-5% of full scale (25W for your slug) or +/- 1.25 Watts,
so the above is within spec. If you want better accuracy, consider the
LP-100A which is +/- 5% of reading. At 12 Watts, this would be +/- 0.6
Watts or twice as good as a Bird...and probably cheaper as well considering
the cost of a comparable set of slugs.
http://www.telepostinc.com/lp100.html
73, Bill
--
View this message in context: http://elecraft.365791.n2.nabble.com/K3-RF-Calibration-RF-output-tp7367721p7368184.html
Sent from the Elecraft mailing list archive at Nabble.com.