[Elecraft] P3 level accuracy
Don Wilhelm
w3fpr at embarqmail.com
Mon Sep 27 18:10:12 EDT 2010
Joe,
Comparing the K3 S-meter with the P3 amplitude indication has some
limitations.
You stated that your K3 has been calibrated with the XG2 and I have no
reason to doubt that, but do realize the granularity in that calibration.
If you calibrate the K3 S-9 indicator bar to just barely come on at -73
dBm, and it will be the only bar illuminated until the signal goes up to
-67 dBm (assuming 6 dB per S-unit). So your indication is well within
that granular step. OTOH, if you calibrated the K3 S-meter so the S-9
indication is on full at -73 dBm, then your K3's S-9 could be as low as
-79 dBm.
When comparing the K3 S-meter reading to the P3, one must consider this
granularity in the K3 S-meter - in other words, it is not an "exact
science" - allow for a +/- 6 dB range when making comparisons unless you
have calibrated the K3 S-meter with this granularity in mind and can
define the S-9 range more closely with accuracy.
I believe Alan is working on additional calibration processes for the
P3, so if we could give this subject a bit of a rest until he completes
that work, we will keep the reflector more quiet for now - after the P3
calibrations are in place, then I am certain comments on accuracies and
inaccuracies of the response will be well received and appreciated.
What we have currently is a digital display with its digital granularity
being compared to an analog display with greater resolution than the
digital display - to me, that is like comparing apples and oranges
unless the granularity (and its consequences) is understood. Those of
us old-timers who grew up in the analog meter world do have a difficult
time with this fact - it just does not seem "natural" unless we have
gone through the "digital indoctrination" and learned its consequences.
73,
Don W3FPR
On 9/27/2010 5:26 PM, Joe Subich, W4TV wrote:
>
> > I've had several replies, all report the XG-2's -73dBm level to be
> > between about -78dBm and -80dBm on the P3, so it's at least
> > consistent, and will, no doubt receive Elecraft's attention in
> > due course!
>
> Well, my XG-2 shows up at -68 to -70 dBm on my P3. The P3 shows
> a no signal noise floor of -150dBm (RX Ant) in a 2 KHz span. The
> K3 has been calibrated to this same XG-2 using the K3 Utility and
> reads S9. The 1 uV output shows up at -105dBm with the same test
> conditions.
>
> The K3 has the IF output modification fitted.
>
> 73,
>
> ... Joe, W4TV
>
More information about the Elecraft
mailing list