Fred,
Your statement of connecting to an equipotential plane is a good one,
but your concern about spikes and inductive reactance back to the
"planet" should not be a concern. It IS important that all items in the
protected area be connected to the SAME point on this safety ground.
Connecting one ESD prevention device to one green-wire ground, and
connecting yet another device to yet another receptacle's green wire
ground will cause a difference in potential between those two points on
the green-wire ground.
The same principles apply to installations like commercial broadcast
stations which must remain in operation throughout a lightning storm.
The solution is to keep everything in the station at the same potential
through the use of a "grounding window" - which is the single point to
which *everything* in the station is connected. In the event of a
strike, it does not matter whether that grounding point raises to
several thousand volts - everything connected to that point will also
rise in potential by that same amount. It is the potential *difference*
of those items that are connected to that single point that is kept to a
minimum. It takes a potential *difference* to cause harm or damage.
It is similar with ESD protection, but on a much smaller scale. Do not
connect your wrist strap to one receptacle ground screw and your
anti-static mat to yet another. Connect them all to the same point.
Failure to do so destroys the design of your anti-static setup. And
that is the reason that many anti-static mats have a connector where you
can attach your wrist strap - use it to keep everything at the same
potential.
73,
Don W3FPR
On 12/18/2010 6:19 PM, Fred Jensen wrote:
> Connecting your equipotential plane to the planet is a mixed bag. If
> the connection is direct and big, it's probably fine. If other things
> use it and might put significant current spikes into it, you might want
> to reconsider. It's sometimes surprising the amount of noise pulses
> that can wander around on the "green wire" in a normal residence. If
> your house is wired to code, it's connected to the neutral and the
> planet at the main electrical entrance [only], which could mean quite a
> bit of wire to your shack and lots of inductive reactance in it for very
> fast pulses.
>