[Elecraft] KX1 SIG Menu Item
Don Wilhelm
w3fpr at earthlink.net
Wed Apr 11 07:41:44 EDT 2007
Joel,
You first must set the 'DISPLAY' to the S-meter mode - then:
The Lnn parameter is set for no bars with no signal (antenna
disconnected), and the Hn parameter is set to adjust the high level end.
If you have a signal generator with a 50 uV output level (like the
Elecraft XG1 or XG2), you can set the indication to S-9 with that 50 uV
input. The Lnn parameter must be set first. The numbers you see are
the default values or the previous settings.
If you still have problems understanding the sequence, please ask again.
73,
Don W3FPR
Joel Weder wrote:
> I've been trying to figure out how the heck to use the SIG menu item in my
> KX1. There's virtually no description of how to set it up, other than the
> cryptic notes in the Menu Functions section on page 65. Can anyone explain
> what the parameters (eg) H5 and L84 are supposed to mean, and exactly what
> procedure should be used to set the S-meter?
>