Well,
I re-aligned it with the swamping network and checked the IF gain with a
known/calibrated 50 micro volt source. This time I know for sure that the
S-Meter reads 25 db over S-9 (65 db) at 50 micro volts with the IF gain set pot
adjusted as low as it will go. There appears to be no ALC action at 1 micro volt
and no distortion of strong signals (excluding my neighbor across street).
Nothing really stood out with voltage and resistance checks. I guess I'll have
to make more through voltage/resistance checks and try tube swaps. While I
was trying to find the source of my outlandish S-Meter readings, a couple of
other problems cropped up. When the calibrator is oscillating or I put the TX
in synch mode, the audio starts to crackle (maybe a noisy tube?). The second
problem is the PTO. Every so often, it suddenly jumps, at random, 1.5 to 2
kc's up or down. Otherwise it is rock stable. I replaced the 1N34A USB/LSB
shift diode, hoping that might be the problem. No joy. Next step is to scrounge
some 1% or 2% silver mica's, some good quality .02 bypass caps and shotgun
the PTO (I'm getting tired of pulling it out). I suspect it might be the 200 pf
"postage stamp" mica, most likely suspect, I have seen plenty of them go bad.
Thanks for reading my long winded dissertation and 73! DE N5MCJ, Jerry
-----------------------------------------------------------------
Thanks to all that replied!
Tried it terminated into 50 ohms, no change. I had to drop the signal to 10
micro volts to get S-9. URM-25 might be way out of calibration, I'll try it
with the HP 606B. Even though, RX is still way too hot. Just about every
signal pins the meter. Maybe it's over coupled. I did not use the "swamping
network" (1K res. and .01 cap.) when I aligned it. All tubes are correct, no mods,
all paper and electrolytic caps had been replaced a few years back. I tested
all the tubes (mostly looking for gas/shorts/leakage) and found a couple that
had some heater to cathode/plate leakage, replaced them and the signal
levels went even higher. Bias seems to be working OK (I can cut RF gain via the
front panel control, re-zero the meter pot and make signal levels more or less
normal). I'll give it another shot with a better sig gen and use the swamping
network.
Thanks again for all the info and 73! DE N5MCJ, Jerry