[ARC5] Visual Alignment - another discovery.
Kenneth G. Gordon
kgordon2006 at frontier.com
Fri May 24 17:37:45 EDT 2013
Taking Hutch's advice (and my own thoughts), I replaced the 12SR7 tube
and its diode detector with a soldered-in 1N34A diode, then connected my
regular o'scope probe to the diode load resistor (R-18 on the BC-454
schematic).
The 1N34A is simply replacing the internal diode of the 12SR7 tube, which I
have removed.
The result is shown here:
http://www.w7ekb.com/Personal/BC454/BC454IF-5.jpg
Compare that with this photo:
http://www.w7ekb.com/Personal/BC454/BC454IF-3.jpg
Conditions of the setup of the sweeper, signal generator, o'scope, input to
the receiver, and all other settings are identical.
The differences are:
1) the probes used (Photo 5 standard o'scope probe, Photo 3 Heathkit RF
probe),
2) the existence, or not, of the diode detector (Photo 5 WITH 1N34A
detector, Photo 3 with NO detector),
3) where the probe is connected (Photo 3 probe connected to that
connection point on the last IF can to which the detector is normally
connected, Photo 5 probe connected to the output of the diode detector and
across the diode load resistor).
It looks to me as though the 12SR7 tube I had been using has some sort of
very serious problem, possibly an internal element to some other element
short, possibly the cathode to filament.
I am next going to use the Heathkit probe and will connect it across the diode
load resistor to see if this elminates the "filled in" trace leaving only the
outline.
However, the "filled in" trace does include other information that the bare
outline trace does not. As to whether or not that information is of any real
value remains to be figured out.
Anyone care to look the photos over and give me their opinions, please?
Kenneth G. Gordon W7EKB