[AMRadio] Is there a doctor in the house?
w5jo at brightok.net
w5jo at brightok.net
Tue Jul 24 17:01:21 EDT 2018
Don't worry about traffic here Charlie. That would be an interesting topic
to see discussed
Jim
W5JO
Moderator
-----Original Message-----
This is not a AM related question, and have also posted this on the CCA
site. Hopefully one of you AM restorers on this site made a living
repairing test equipment, and instruments - like meter movements. I have a
45 year old Simpson 260 that has become inaccurate, and it is not the
electronics of the meter. The meter movement is a 50ua and 250MV full scale
unit. When a known test voltage is put onto the meter terminals with no
other connections, the meter is a constant 30% off to the low side, across
the scale. Inject 71MV, get a 50 indication on the meter movement, inject
357MV, you get full scale 250MV. Under a magnifier, the meter is in
excellent mechanical shape, and the needle moves without binding. So, to an
instrument guy, what happened? Please reply off list to keep the traffic to
a minimum. Thanks.
Charlie, W4MEC in NC
More information about the AMRadio