[AMRadio] [Glowbugs] Re: [AWA] Tek 535/545 vertical amp calibration question
Larry Szendrei
ne1s at securespeed.us
Wed Nov 1 07:25:36 EDT 2017
On 10/28/17 12:26 PM, Michael Hanz wrote:
> Just make one - it seems pretty simple:
> https://groups.io/g/TekScopes/topic/951626?p=,,,20,0,0,0::,,,0,0,0,951626
>
> On 10/28/2017 9:41 AM, Larry Szendrei ne1s at securespeed.us
> [antiquewirelessassociation] wrote:
>>
>> I wonder if anyone has any tips for how to calibrate the vertical
>> amplifier on a Tektronics 535 and/or 545 oscilloscope without the EP53A
>> Gain Set Adaptor called for in the manuals' procedure?
>>
A final note on this topic that may prove helpful to anyone who needs to
calibrate the vertical amp on one of these fine old, but beastly,
instruments:
Using the information in the link that Mike provided (see above), I was
able to take one channel of a spare, unused type CA dual-trace plug-in I
had and convert it to calibrate the vertical amps in both my 535A and
RM45 'scopes. One input of the plug-in now bypasses its amplifier and
accepts a signal from the internal calibrator to set the sensitivity of
the mainframe vertical amplifier to 100mv/cm. After calibrating the
mainframe amp I calibrated the plug-ins (two more type CAs and one type
K). After (and despite) the repair/maintenance dose to the scopes, which
in the case of the RM45 was quite extensive, the calibrations weren't
that far off.
73,
-Larry/NE1S